Hi Rellab Guy,
It’s not easy being confronted by something like this, is it? First off, you need to ensure that your PCT system was not the one that caused this. Maybe this is the point being raised by FA. If you look at Jedec JESD22-A102, it says that the ionic contam of the test apparatus and chamber must be controlled to avoid artifacts. It also specifies a minimum resistivity for the water to be used.
You must show FA that you do not have any problems with the DI water you’re using for PCT before you go further with root cause analysis. Other signs that indicate you may indeed have a water problem: 1) other lots loaded with the lot also show metal corrosion or discoloration; 2) other PCT chambers are affected; 3) even the good units show metal discoloration; 4) the affected units do not show passivation defects; and 5) your D/I water has a Cl content of 0.2 ppm or higher.
Keep us posted!